    On 04 May 2016, at 15:37, Cochran, Malcolm wrote:

    Dear Dr. Lelièvre-Berna,
    I am the Scientific Associate supporting the NSE at SNS, Beam Line 15. We have received a proposal that requires a temperature of 900 C. Our First Instrument Scientist, Piotr Zolnierczuk, contacted Peter Falus to inquire about the highest temperature they could achieve at their beam line and received the response seen below. I asked Chris Redmon about purchasing and winding such bifilar heater wires and the best way to place them in a sample environment. He suggested I contact you, as you likely led the development project. Would you provide me with information about the company from which this heater was purchased, please? Also, anything you could share with me to assist me in building such a thing would be greatly appreciated.

    Thank you sincerely,
    Malcolm J. Cochran
    Neutron Spin Echo (BL 15) Engineer
    Building 8600 Room B-419
    MS 6473
    Cell: 865-368-0814
    Office: 865-241-1949

    #2127

    Dear Malcolm,

    Sorry for my late reply, we were collecting infos from IN11 because that furnace is directly managed by the instrument team. This furnace allows to keep the sample in a vacuum or in gas.

    To summarise, the core of the furnace is made from a chamber (the smallest one shown on the photo) with 2 Thermocoax heaters wound at the top and at the bottom. Each Thermocoax is made of 2 wires wound directly in a groove machined on the chamber. As these two wires are wound together, the current circulates in both ways and there is no generation of magnetic field.

    This chamber does not allow to heat above 450°C (700K). Thermocoax heaters can be used at up to 1000°C but the conditions in which they are used prevents the exploitation of higher temperatures. A solution would be to brase them homogeneously.

    I do not have a quotation of the heater used in this furnace so I provide another quote we requested to build compact heaters for cradles. The current also circulates in both ways and the sample is glued at the extremity of a cartridge containing that heater. With this solution, the Thermocoax heater is replaceable and the sample is heated up to 800°C in vacuum.

    Hope this helps,
    Eddy